Updated 6 July 2026
  • FOC is a dual-context concept that maps observed microcavity filling and overflowing regimes in lab-on-a-chip experiments and informs large-scale liquid handling in JUNO.
  • In microfluidics, FOC delineates filling versus overflowing states by examining interface pinning, capillary forces, and transient dynamics using advanced imaging and simulations.
  • In JUNO, the FOC system integrates precise detector filling, overflow buffering, and optional circulation to meet stringent radiopurity and hydrostatic control requirements.

Searching arXiv for the cited FOC-related papers to ground the article in current sources. {"query":"arXiv (Nagel et al., 2024) Experimental and Numerical Study of Microcavity Filling Regimes for Lab-on-a-Chip Applications", "max_results": 5} Searching arXiv for the microcavity filling paper. {"query":"Experimental and Numerical Study of Microcavity Filling Regimes for Lab-on-a-Chip Applications", "max_results": 10} Filling, Overflowing, and Circulating System (FOC) is a designation used in recent literature for two distinct but thematically related technical objects. In microfluidics, it serves as a mapping from observed microcavity-interface states to the labels Filling and Overflowing, while a distinct Circulating regime was not identified under the low-ReRe, capillary-dominated conditions studied (Nagel et al., 2024). In the Jiangmen Underground Neutrino Observatory (JUNO), FOC denotes the liquid-handling subsystem responsible for detector filling, overflow management, and optional circulation for re-purification of liquid scintillator (LS) and water (Li et al., 12 Jul 2025, Li et al., 16 Dec 2025). In both contexts, the term centers on controlled management of interfaces, inventories, and transients under strong geometric and safety constraints.

1. Terminological scope and domain-specific meanings

The acronym is not used uniformly across domains. The microcavity study defines experimentally observed filling states and then maps them onto the Filling, Overflowing, and Circulating notion, with explicit caution that sustained in-cavity recirculation eddies were not observed as a regime (Nagel et al., 2024). By contrast, JUNO uses FOC as the name of a process plant and control system; the nomenclature varies slightly between publications, appearing as “Filling, Overflowing, and Circulating System” and as “Filling, Overflow, and Circulation (FOC) system” (Li et al., 12 Jul 2025, Li et al., 16 Dec 2025).

Usage Filling / Overflowing Circulating
Microcavity flow Filling: cavity invasion after pinning; Overflowing: interface skims over cavities leaving an air cushion or bubble Not identified as a distinct regime
JUNO liquid handling Filling: water fill and LS exchange; Overflowing: passive and active accommodation of thermal volume changes Optional online extraction, purification, and reinjection

In the microcavity context, Filling means that all cavities are invaded, either voidlessly or with only tiny trapped bubbles at the cavity bottom. Overflowing means that the interface pins at cavity edges, detaches, and passes over the cavities without complete filling, leaving an air cushion or bubble below the liquid. The paper also defines a Partially Filled state, but reports that it is not repeatable across replicates (Nagel et al., 2024).

In JUNO, Filling refers to commissioning operations such as synchronous water filling of the Central Detector (CD) and Water Pool (WP), and later water-to-LS exchange. Overflowing refers to the buffering of LS thermal volume changes via passive overflow tanks and active transfers with a storage tank. Circulation refers to optional extraction from the detector bottom, external re-purification, and reinjection at the top (Li et al., 16 Dec 2025).

2. Microcavity FOC regimes in capillary-dominated flow

The microcavity study investigates a moving liquid-air interface over a 3×33\times 3 array of cylindrical cavities in a rectangular microchannel under lab-on-a-chip conditions. Two cavity depths were tested, d=0.07 mmd = 0.07\ \mathrm{mm} and d=0.14 mmd = 0.14\ \mathrm{mm}. The cavities were patterned into a PMMA slide with micromilled roughness ≤5×10−6 m\le 5\times 10^{-6}\ \mathrm{m}, sealed by a smooth transparent Zeonex lid. Top-down high-speed imaging with 3 μm3\ \mu\mathrm{m} pixel size and grayscale-based edge detection was used to track the interface (Nagel et al., 2024).

Three fluids were examined at 20∘C20^\circ\mathrm{C}: water, a 0.1%0.1\% aqueous Tween-80 solution, and Novec-7500. Their wetting properties differ strongly. Water and Tween exhibit partial wetting on PMMA/Zeonex, whereas Novec is reported as perfectly wetting on PMMA and Zeonex with θeq=0∘\theta_{eq}=0^\circ. Static Zeonex contact angles from droplet tests were 94±5∘94\pm 5^\circ for water, 85±5∘85\pm 5^\circ for Tween, and 0∘0^\circ for Novec. The experimentally achieved capillary-number ranges were 1.37×10−61.37\times 10^{-6}, 1.37×10−51.37\times 10^{-5}, and 1.37×10−41.37\times 10^{-4} for water; 2.85×10−62.85\times 10^{-6}, 2.85×10−52.85\times 10^{-5}, and 2.85×10−42.85\times 10^{-4} for Tween; and 7.67×10−47.67\times 10^{-4} and 7.67×10−37.67\times 10^{-3} for Novec (Nagel et al., 2024).

The regime logic is governed by pinning, depinning, wetting, and viscous forcing. Filling occurs when the interface pins at upstream cavity edges and subsequently advances through the cavity volume. The mechanism is described as low apparent dynamic contact angle θd\theta_d, strong capillary suction relative to imposed viscous drag, and pinning that precedes invasion. Near-wall cavities often invade first because of rivulets and local curvature, followed by center cavities. Overflowing also begins with pinning, but pinning is then insufficient to promote snap-through; the interface detaches and skims over the cavities, leaving an air cushion or bubble. Partial filling is attributed to sensitivity to local roughness, transient surfactant distribution, and subtle contact-angle hysteresis, producing stochastic pinning and depinning (Nagel et al., 2024).

The principal parametric trends are discrete rather than threshold-fitted. Novec exhibits Filling across all tested capillary numbers and both tested depths. Tween depends on capillary number and depth: at Ca=2.85×10−4Ca = 2.85\times 10^{-4}, Overflowing occurs, while lower CaCa and shallower cavities increase the probability of Filling. Water is predominantly Overflowing; Partial filling was observed only at Ca=1.37×10−6Ca = 1.37\times 10^{-6} in shallow cavities, while all higher tested CaCa values yielded Overflowing. The study reports that apparent dynamic contact angles measured upstream of the array increase by more than 10∘10^\circ across the tested CaCa range for each fluid, and that deeper cavities are harder to fill (Nagel et al., 2024).

The relevant dimensionless and capillary relations are given explicitly. The capillary number is

Ca=μUσ,Ca = \frac{\mu U}{\sigma},

the Reynolds number is

Re=ρULμ,Re = \frac{\rho U L}{\mu},

the Bond number is

Bo=ρgL2σ,Bo = \frac{\rho g L^2}{\sigma},

and the Young-Laplace capillary pressure is

Δp=σ(1R1+1R2).\Delta p = \sigma \left( \frac{1}{R_1} + \frac{1}{R_2} \right).

A Cox-Voinov-type relation is cited as a common reference model,

θd3=θe3+9 Ca ln⁡ ⁣(Lλ),\theta_d^3 = \theta_e^3 + 9\,\mathrm{Ca}\,\ln\!\left(\frac{L}{\lambda}\right),

but the study does not fit a specific θd(Ca)\theta_d(Ca) law (Nagel et al., 2024).

3. Interface transients, numerical reconstruction, and the non-observation of a microcavity circulating regime

The microcavity work combines experiments with $3$D Volume-of-Fluid simulations using plicRDF-isoAdvection on unstructured meshes. The solver is interFlow in OpenFOAM v2212 within the TwoPhaseFlow framework. A baseline mesh resolution of Δx=0.033 mm\Delta x = 0.033\ \mathrm{mm} was used, with refinement tests at 0.02 mm0.02\ \mathrm{mm} and 0.01 mm0.01\ \mathrm{mm}; these showed negligible change in interface progression relative to regime differences. The timestep obeyed a capillary-wave restriction,

Δt<(ρ−+ρ+)(Δx)34πσ.\Delta t < \sqrt{\dfrac{(\rho^- + \rho^+) (\Delta x)^3 }{4 \pi \sigma} }.

Boundary conditions were constant inlet velocity and zero-pressure outlet, with partial-slip walls of slip length λ=0.02 mm\lambda = 0.02\ \mathrm{mm}, an xzxz symmetry plane, and constant wall contact angles taken from measured θd\theta_d medians. Stabilization used artificial interface viscosity μΣ=μ\mu_\Sigma = \mu and wisp filtering tolerance ϵw=10−5\epsilon_w = 10^{-5} (Nagel et al., 2024).

The transient behavior clarifies how the observed regimes unfold. In a Novec Filling case at Ca=7.67×10−4Ca = 7.67\times 10^{-4}, the reported sequence is pinning at first-row edges at t≈0.16 st \approx 0.16\ \mathrm{s}, rapid filling of near-wall cavities over Δt≈0.015 s\Delta t \approx 0.015\ \mathrm{s} at t≈0.175 st \approx 0.175\ \mathrm{s}, strong pinning at the center cavity at t≈0.19 st \approx 0.19\ \mathrm{s}, and snap-through with traversal at t≈0.20 st \approx 0.20\ \mathrm{s}. Position-versus-time curves exhibit “kinks” that coincide with pinning and filling events. In a Tween Overflowing case at Ca=2.85×10−4Ca = 2.85\times 10^{-4}, the first row shows pinning at t≈0.18 st \approx 0.18\ \mathrm{s}, center-cavity detachment at t≈0.20 st \approx 0.20\ \mathrm{s}, and wall-adjacent detachment at t≈0.22 st \approx 0.22\ \mathrm{s}, followed by repetition at the second row around t≈0.24 st \approx 0.24\ \mathrm{s} (Nagel et al., 2024).

Agreement between experiments and simulations is reported as strong for most Filling-versus-Overflowing cases and for the timing of pinning, detachment, and traversal kinks. Partial filling was not reproduced; in marginal cases, simulations tended to predict Overflowing. The authors characterize this as conservative for void-free design. Near-wall rivulet morphology remains harder to reproduce exactly, and for Novec the effective contact-angle calibration depended on capillary number: θ=45∘\theta = 45^\circ at Ca=7.67×10−3Ca = 7.67\times 10^{-3} and θ≈28∘\theta \approx 28^\circ at Ca=7.67×10−4Ca = 7.67\times 10^{-4} gave the best matches among tested values (Nagel et al., 2024).

The status of “Circulating” in this microfluidic usage is therefore negative rather than positive: it is not identified as a distinct regime. No sustained in-cavity recirculation eddies are reported by either experiments or simulations under the tested low-ReRe conditions. The imaging is top-down, intra-cavity flow fields were not resolved, and the modeling focuses on interface pinning, invasion, and skimming rather than on cavity-resolved vortical structures. The paper explicitly states that design implications for inducing or suppressing recirculation are extrapolations and should be tested separately. A plausible implication is that, in this parameter space, interfacial wetting barriers dominate more strongly than intra-cavity inertial recirculation (Nagel et al., 2024).

For practical diagnosis, the study recommends observing the pinning-depinning sequence at cavity edges, identifying interface kinks in position-time curves when rows are crossed, measuring upstream θd\theta_d, and using simulations with calibrated θd\theta_d to predict regime and transients. The associated data and code are publicly available through Bosch Research GitHub and the TUDatalib repository (Nagel et al., 2024).

4. JUNO FOC as a large-scale detector liquid-handling and control subsystem

In JUNO, the FOC system is a plant-scale subsystem engineered to safely deliver, maintain, and re-purify the $20$-kiloton LS inventory of the Central Detector while preserving 3%3\% energy resolution at 1 MeV1\ \mathrm{MeV} and low radioactive background. The control objectives include protection of the $35.4$-m-diameter acrylic sphere and its chimneys during filling and exchange, stable detector internal pressure during operation, liquid-level excursions limited to ≤20 cm\le 20\ \mathrm{cm}, automatic overflow and refill, and online LS circulation for re-purification. Purity constraints include avoidance of oxygen, moisture, dust, radon, and rare-gas ingress, leakage rate ≤10−6 mbar⋅L⋅s−1\le 10^{-6}\ \mathrm{mbar\cdot L\cdot s^{-1}}, optical attenuation length >20 m>20\ \mathrm{m}, and dust <10 mg<10\ \mathrm{mg}. Reliability targets include sensor accuracy typically ≤0.2%\le 0.2\% FS, closed-loop deviation ≤1%\le 1\%, pump speed stability within ±0.2%\pm 0.2\% of setpoint, on-off valve response <1 s<1\ \mathrm{s}, regulating-valve positioning accuracy ±1%\pm 1\%, and extensive redundancy with fail-safe behavior (Li et al., 12 Jul 2025).

The hardware core is a Siemens S7-300 PLC with one CPU and one interface module, six $8$-channel AI modules, two $8$-channel AO modules, and three digital I/O modules. The PLC cabinet includes an industrial audible/visual alarm device and a physical emergency stop. Instrumentation is dominated by Endress+Hauser devices. For the CD, the system includes four electronic remote differential-pressure level gauges in two redundant pairs at 0.20%0.20\% FS, one laser level gauge at 0.10%0.10\% FS, and two differential-pressure gauges on the upper chimney at 0.20%0.20\% FS. WP level uses five static pressure level gauges at 0.20%0.20\% FS. Tank instrumentation combines differential-pressure gauges and radar wave level gauges at 0.10%−0.20%0.10\%-0.20\% FS. Flow is measured by Coriolis and vortex meters, and pressure by gas-space and pump-discharge transmitters (Li et al., 12 Jul 2025).

Actuation is provided by SED regulating and on-off valves and by CDR electromagnetic pumps together with Yamei vacuum self-priming pumps. Eight pump units are included in total. Electromagnetic pumps are configured one-operating/one-standby. The process also contains a storage tank, two overflow tanks, top and bottom chimneys, a nitrogen cover-gas system, and filtration monitored by differential pressure. The software stack is built in Siemens TIA Portal and implements PID control, sequential or state-machine control, interlocks, split-range control, and selective control. EPICS is used for live monitoring, historical queries, and data caching across JUNO subsystems, with a heartbeat program supervising PLC status. Alarm management spans cabinet sirens, HMI alarms, and emergency stop, while historian functionality uses dual storage with local disk and synchronized Detector Control System (DCS) storage (Li et al., 12 Jul 2025).

The control methods are described in operational rather than formal-model terms. Ziegler-Nichols rules supply initial KpK_p, TiT_i, and TdT_d, followed by empirical refinement for minimal overshoot and fast settling. Signal preprocessing includes arithmetic mean filtering, which reduced radar level-gauge noise from about 30 mm30\ \mathrm{mm} to about 5 mm5\ \mathrm{mm}, and time-hysteresis filtering to reduce transient false trips. Split-range control is used, for example, when a single flow controller drives two regulating valves during water filling. Selective control prioritizes the most critical constraint, such as pressure or level, when limits are approached. Safety logic is independent of normal control; critical sensors are separated in safety interlocks, and fail-safe defaults close valves and stop pumps upon anomaly detection or signal loss (Li et al., 12 Jul 2025).

5. Filling, overflow, and circulation operations in JUNO

The FOC system spans commissioning and long-term operation. During commissioning, water filling synchronously fills the CD and WP to displace air, rinse the acrylic inner surface, and establish a safe hydrostatic environment before LS exchange. LS filling then injects purified LS from the top chimney while draining water from the bottom. During long-term operation, overflow tanks buffer thermal volume changes and circulation supports optional online purification (Li et al., 16 Dec 2025).

Water filling handled approximately 6.4×104 m36.4\times 10^{4}\ \mathrm{m^3} in $45$ days. The reported average flow calculation is

Q=6.4×104 m345×24×3600 s=0.0165 m3/s=16.5 L/s=59.3 m3/h.Q = \frac{6.4\times 10^{4}\ \mathrm{m^3}}{45\times 24\times 3600\ \mathrm{s}} = 0.0165\ \mathrm{m^3/s} = 16.5\ \mathrm{L/s} = 59.3\ \mathrm{m^3/h}.

Operationally, the total available flow was about 90 m3/h90\ \mathrm{m^3/h}, with Stage-2 and Stage-3 setpoints around 80 m3/h80\ \mathrm{m^3/h}, punctuated by daily pauses and slow on-off top-up bursts. The WP served as the primary level reference, and the differential level ΔH=HCD−HWP\Delta H = H_{CD} - H_{WP} was held within a fraction, ≤25%\le 25\%, of the FEA design limit, with on-off interlocks if exceeded. Because the acrylic sphere tolerates higher external than internal pressure, a slightly higher WP level was preferred. Pure-water system tests reported two PID-controlled streams reaching about 39.3 m3/h39.3\ \mathrm{m^3/h} on the WP pipeline and about 18.9 m3/h18.9\ \mathrm{m^3/h} on the CD pipeline in about $2$ minutes, then holding with ≤0.5%\le 0.5\% uncertainty (Li et al., 16 Dec 2025, Li et al., 12 Jul 2025).

The tall risers imposed a vertical static head of about 44 m44\ \mathrm{m}, creating risk of vapor cavity formation and water hammer. Vacuum-break valves on WP lines and a 1500 L1500\ \mathrm{L} nitrogen buffer on CD lines prevented sub-atmospheric pressures without air ingress. A typical purge flow of about 15 m3/h15\ \mathrm{m^3/h} is reported for sensitive operations (Li et al., 16 Dec 2025).

LS filling handled approximately 2.3×104 m32.3\times 10^{4}\ \mathrm{m^3} in $200$ days, corresponding to an average of

Q=2.3×104 m3200×24×3600 s=0.00133 m3/s=1.33 L/s=4.79 m3/h.Q = \frac{2.3\times 10^{4}\ \mathrm{m^3}}{200\times 24\times 3600\ \mathrm{s}} = 0.00133\ \mathrm{m^3/s} = 1.33\ \mathrm{L/s} = 4.79\ \mathrm{m^3/h}.

The nominal LS production rate was 7 m3/h7\ \mathrm{m^3/h}. Before exchange, the CD was purged with ultra-pure nitrogen at 222Rn<10 μBq/m3^{222}\mathrm{Rn} < 10\ \mu\mathrm{Bq/m^3}. Startup injected an initial 100 m3100\ \mathrm{m^3} LS batch and held for several days for OSIRIS radiopurity verification and radon decay stabilization. Continuous filling then ramped from $1$-2 m3/h2\ \mathrm{m^3/h} to 7 m3/h7\ \mathrm{m^3/h}, with top injection and bottom water drainage. After completion, about 80 m380\ \mathrm{m^3} was drained from the CD bottom to remove the earliest batch, which had the longest contact with residual water and particulates. Internal FOC tests also demonstrated LS flow regulation to 1500 L/h1500\ \mathrm{L/h} through a regulating valve, Coriolis meter, and filter with PID feedback, reaching ±0.5%\pm 0.5\% accuracy after about $3$ minutes (Li et al., 16 Dec 2025, Li et al., 12 Jul 2025).

Hydrostatic management is central during LS exchange because ρLS≈0.856 g/cm3\rho_{LS} \approx 0.856\ \mathrm{g/cm^3} and ρw≈1.0 g/cm3\rho_w \approx 1.0\ \mathrm{g/cm^3}. The governing hydrostatic relation is

P=ρgh.P = \rho g h.

To balance the denser external water, the LS free surface must be higher than the WP level. The reported illustrative balance is

ρLSghLS≈ρwghWP,\rho_{LS} g h_{LS} \approx \rho_w g h_{WP},

giving

hLS/hWP≈ρw/ρLS≈1.169.h_{LS}/h_{WP} \approx \rho_w/\rho_{LS} \approx 1.169.

For hWP≈40 mh_{WP} \approx 40\ \mathrm{m}, this implies hLS≈46.8 mh_{LS} \approx 46.8\ \mathrm{m}, which is consistent with the final LS level near 47 m47\ \mathrm{m}. Achieved LS level precision was ±2 cm\pm 2\ \mathrm{cm}, and flow regulation was within ±0.5%\pm 0.5\% of setpoints (Li et al., 16 Dec 2025).

Overflow management combines passive and active modes. Two 50 m350\ \mathrm{m^3} overflow tanks connected to the top chimney absorb small thermal swings by gravity flow. For larger temperature steps, transfers occur between storage and overflow or CD via pumps under PLC control. With the operational CD level stabilized at about 47 m47\ \mathrm{m}, passive overflow capacity is reduced to about 80%80\% of nominal, so the PLC maintains CD level within ±5 cm\pm 5\ \mathrm{cm} through small active transfers in addition to passive buffering. The thermal-volume estimate

ΔV=βVΔT\Delta V = \beta V \Delta T

with β≈8.8×10−4/∘C\beta \approx 8.8\times 10^{-4}/^\circ\mathrm{C} and V≈2.3×104 m3V \approx 2.3\times 10^{4}\ \mathrm{m^3} yields ΔV≈20.2 m3/∘C\Delta V \approx 20.2\ \mathrm{m^3}/^\circ\mathrm{C}, which is consistent with the reported buffering demand (Li et al., 16 Dec 2025).

Circulation extracts LS from the detector bottom, routes it through water extraction and gas stripping, and reinjects it at the top. The nominal circulation setpoint is 7 m3/h7\ \mathrm{m^3/h}, corresponding to a nominal full-volume turnover time of about $137$ days. CFD results cited in the operational paper indicate that a single-volume exchange removes about 37%37\% of bulk impurities without temperature offset, while heating the injected LS by about 5∘C5^\circ\mathrm{C} improves circulation efficiency to about 55%55\% by suppressing convective short-circuiting (Li et al., 16 Dec 2025).

6. Purity assurance, operational limitations, and broader significance

For JUNO, FOC performance is inseparable from radiopurity and cleanliness control. All wetted components are specified as low-radioactivity stainless steel with U<0.7 ppbU < 0.7\ \mathrm{ppb}, electropolished inner roughness ≤0.4 μm\le 0.4\ \mu\mathrm{m} and typically $0.2$-0.3 μm0.3\ \mu\mathrm{m}, helium leak-rate goals of <10−8 mbar⋅L⋅s−1<10^{-8}\ \mathrm{mbar\cdot L\cdot s^{-1}} for components and <10−6 mbar⋅L⋅s−1<10^{-6}\ \mathrm{mbar\cdot L\cdot s^{-1}} for assembled systems or critical flanges, and nitrogen inerting over LS-contacted gas spaces. Cleaning is specified as detergent degrease, dilute HNO3\mathrm{HNO_3} pickling, HNO3\mathrm{HNO_3} passivation, UPW rinse, and nitrogen drying, with acceptance by MIL-STD-1246C particle-density criteria and final-rinse U/Th<10−14 g/gU/Th < 10^{-14}\ \mathrm{g/g} (Li et al., 16 Dec 2025).

The achieved radiopurity metrics during LS filling were 222Rn<1 mBq/m3^{222}\mathrm{Rn} < 1\ \mathrm{mBq/m^3}, below the stated requirement of <5 mBq/m3< 5\ \mathrm{mBq/m^3}, and bulk 238U/232Th<10−16 g/g^{238}\mathrm{U}/^{232}\mathrm{Th} < 10^{-16}\ \mathrm{g/g} by ICP-MS. An initial 210Po^{210}\mathrm{Po} rate of about 5×104 cpd/kton5\times 10^{4}\ \mathrm{cpd/kton} after fill is reported. Monitoring used in-situ 214Bi^{214}\mathrm{Bi}-214Po^{214}\mathrm{Po} tagging for radon, periodic $3$-5 t5\ \mathrm{t} OSIRIS batch assays, and ICP-MS on prepared LS samples with ultra-clean labware. The radon-decay relation

N(t)=N0e−λt,λ=ln⁡2/T1/2,N(t) = N_0 e^{-\lambda t}, \quad \lambda = \ln 2 / T_{1/2},

with T1/2(222Rn)≈3.82 dT_{1/2}(^{222}\mathrm{Rn}) \approx 3.82\ \mathrm{d}, is used to illustrate why the initial 100 m3100\ \mathrm{m^3} hold stabilizes nonequilibrium radon before continuous fill (Li et al., 16 Dec 2025).

Both domains also report limitations. In the microcavity study, top-down imaging does not resolve intra-cavity flow fields, surfactant transport is not modeled, the dynamic contact angle is imposed as a constant apparent θ\theta rather than through a fitted θd(Ca)\theta_d(Ca) law, and stochastic Partial filling linked to roughness and adsorption is not reproduced numerically. The absence of a distinct Circulating regime is therefore an observation specific to the studied low-ReRe, capillary-dominated parameter range rather than a universal exclusion of cavity recirculation (Nagel et al., 2024).

In JUNO, instrumentation tests revealed a $0$-0.3 m0.3\ \mathrm{m} dead zone for some differential-pressure level gauges due to installation constraints and about a 0.2 m0.2\ \mathrm{m} dead zone with response lag for radar gauges in horizontal tanks. An overflow-line airlock was encountered after filling and was mitigated by repurposing the LS fill line, sustaining overflow above 2 m3/h2\ \mathrm{m^3/h}. Raw LAB supply intermittency and scheduled maintenance caused short pauses during LS filling. Structural loads remained within design, although a few rods reached about 10%10\% above FEA predictions and were reported as under investigation. ICS cybersecurity specifics, line sizes, and some gas-handling details are not publicly specified in the paper set (Li et al., 12 Jul 2025, Li et al., 16 Dec 2025).

The broader significance differs by scale but converges methodologically. In microfluidics, FOC language organizes how wetting, pinning, and capillary suction determine whether cavity invasion is achieved or bypassed. In JUNO, FOC integrates hydraulics, automation, leak-tight engineering, nitrogen protection, and radiopurity assurance into a subsystem that filled the world’s largest LS detector with verified ±2 cm\pm 2\ \mathrm{cm} level precision and ±0.5%\pm 0.5\% flow regulation during LS operations (Li et al., 16 Dec 2025). JUNO explicitly situates this approach in conceptual continuity with Borexino and KamLAND—nitrogen blanketing, ultra-clean materials, distillation, stripping, water extraction, leak-tight assembly, and online radiopurity assay—while emphasizing the scale-driven advances required for a 35.4 m35.4\ \mathrm{m} sphere, synchronized WP-CD hydrostatic control, and overflow handling for thermal swings of about 20 m3/∘C20\ \mathrm{m^3}/^\circ\mathrm{C} (Li et al., 16 Dec 2025).